Mechanism of Autoclave Sterilization
The autoclave is like a large pressure cooker and is a physical method for disinfection and sterilization. Autoclaves work with a combination of steam, pressure and time. Autoclaving does not remove chemical contamination. Steam must be in contact with the material being sterilized to be effective. Steam is generated and pumped into the chamber raising the temperature and pressure in order to kill microorganisms. Moist heat kills more effectively than dry heat because the mass transfer of heat as steam condenses is very efficient. Therefore, if a load is contained within a bag or other enclosure, then the enclosure must not be tightly sealed (e.g., in the case of surgical instruments) or the enclosure must contain some water to ensure steam production (e.g., waste bags).Source: UC Santa Barbara, Environmental Health and SafetySterilization and Disinfection Parameters
Use a minimum cycle time of 121-124°C (250-255°F), = 30 minutes,15 psi for the disinfection of medical waste. These parameters are required by the California Department of Public Health for the disinfection of medical waste. Items have to be hot enough for long enough in the presence of steam to disrupt protein structure. At 121°C, or 250°F, saturated steam destroys vegetative cells and endospores in a small volume (~100mL) within 10-12 min. Treatment must continue for a total of 30 minutes to provide a margin of safety. Sterilization of larger volumes will require a longer cycle time.